						I am looking for suggestions in analyzing C1-C7 aliphatic acids in a water matrix that has can have % level of ethylene glycol present.   We are looking for detection limits in the range of 10 - 50 ppm.  What type of SPE and or sample prep would be appropriate?   Any column and/or mobile phase suggestions are also welcome.
					
									- 
								
 - Posts: 3594
 - Joined: Mon Aug 30, 2004 11:43 pm
 
						Call Grace/Alltech technical support and ask them.  I've been happy with their Prevail Organic Acid column.
					
									- 
								
 - Posts: 195
 - Joined: Fri Oct 22, 2004 6:02 pm
 
						Dionex offers the Acclaim Organic Acid column for this application. It is a silica based high-efficiency RP column which is fully compatible with highly aqueous mobile phases. The column housing is non-metallic - PEEK, which would minimizemetal contamination from SST column hardware.
The condition for separating C1 to C7 organic acids:
Column: Acclaim OA, 150x4.0 mm
Mobile Phase: A - MeCN, B - 0.05% phosphoric acid or methanesulfonic acid
Gradient: 0 to 45% A in 11 min and keep for 5 minute (minor modification might be needed)
Flow rate: 1 mL/min
Detection: UV at 210 nm
Here is the typical result you should expect (in Figure 7).
http://www1.dionex.com/en-us/webdocs/25 ... OA_V22.pdf
Due to low wavelength detection, the mobile phase needs to be of carefully prepared with high quality reagents.
If ethylene glycol content is less than 5%, I don't expect too much of trouble. If it does impose some difficulties, using a 250x4.0 mm column should help to some extent.
In case these acids can be analyzed by a RP column due to matrix effect, please let me know so that I can offer an alternative.
Good luck!
									The condition for separating C1 to C7 organic acids:
Column: Acclaim OA, 150x4.0 mm
Mobile Phase: A - MeCN, B - 0.05% phosphoric acid or methanesulfonic acid
Gradient: 0 to 45% A in 11 min and keep for 5 minute (minor modification might be needed)
Flow rate: 1 mL/min
Detection: UV at 210 nm
Here is the typical result you should expect (in Figure 7).
http://www1.dionex.com/en-us/webdocs/25 ... OA_V22.pdf
Due to low wavelength detection, the mobile phase needs to be of carefully prepared with high quality reagents.
If ethylene glycol content is less than 5%, I don't expect too much of trouble. If it does impose some difficulties, using a 250x4.0 mm column should help to some extent.
In case these acids can be analyzed by a RP column due to matrix effect, please let me know so that I can offer an alternative.
Good luck!
Xiaodong Liu
				- 
								
 - Posts: 583
 - Joined: Thu Feb 23, 2006 3:15 am
 
						We have customers using Unison UK-C18 for organic acids as well.
(using 0.1% H3PO4 as their mobile phase).
Depending on what you are analyzing, you may have analytes eluting with k'<1.
So you should look for dispersion from your LC (e.g. autosampler, tubing, detector).
Also make sure you sampling rate is high enough on your aquisition software.
									(using 0.1% H3PO4 as their mobile phase).
Depending on what you are analyzing, you may have analytes eluting with k'<1.
So you should look for dispersion from your LC (e.g. autosampler, tubing, detector).
Also make sure you sampling rate is high enough on your aquisition software.
- 
								
 - Posts: 31
 - Joined: Mon Oct 29, 2007 7:45 pm
 
						Thanks for all of the suggestions!!
					
									
			5 posts
							Page 1 of 1
					
	Who is online
				In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 14 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 