In GC-MS library search report, what is the meaning of goting the following in every run?

1- GC stationary phase
2- septum bleed
3- diisooctylphthalate
4- butyl-2-methylpropylphthalate

1 is from your column. 2 is from the septum. 3 and 4 are platicizers which are coming from some plastic that which has been in contact with your sample or the solvents used in processing your sample. They are in every run because they are contaminants present in your samples. You can probably just ignore these peaks unless they become very large or have the same retention time as your analyte.

how can I avoid them?
is 1 - stationary phase will effect the column? its life? its performance?

Use a column with low bleed that is made for GC/MS. Use the best available septum for your GC injection port. Use the highest purity solvents that you can afford.
4 posts Page 1 of 1

Who is online

In total there are 225 users online :: 1 registered, 0 hidden and 224 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Majestic-12 [Bot] and 224 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry