Valproic acid: peak tailing problem

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Dear friends,
Am doing the method development of Valproic acid. IK have problems with the base line (very high ) and the peak shape is bad with severe tailing. I have used almost all cominations olf LCMS buffers as Mp and tried with different columns. The method is a SIM.
If anyone has any idea about the method, please help me

shan

hi

i would ask u first the basic question about valproic acid. it is not fragmenting in LCMSMS. so can u please tell me which ion you have selected?? As we have done valproic acid as PsedoSRM method in LCMSMS.
ADPATEL

One of our customers had same problem with tailing of valproic acid on a regular C18 column. We developed the following method for LC/MS:
http://www.sielc.com/compound_207.html

regards,

Vlad

dear aplesh

Even am using a pseudo SRM for the amu 142.9. Could you please get me some literature which is helpful.

First: What ESI mode are you using? As VA is an acid, you should use Negative Mode...
5 posts Page 1 of 1

Who is online

In total there are 231 users online :: 1 registered, 0 hidden and 230 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Google [Bot] and 230 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry