Hi,

I'm pretty new to the forums and to liquid chromatography in general. I'm running a LCMS method on a C18 column and I was wondering if it is ok to wash and store columns in mobile phase containing acid, in this case 0.1% formic acid.

Thanks!

generally yes. just make sure that there is no salt on the column and that some amount of organic is in the mobile phase you use for storage. 10% acetonitrile with 0.1% FA will do the job.
success
goxy

I would not store the column in an acidic solution. The pH of 0.1% formic acid is about 2.7. This can cause slow hydrolysis of the bonded ligands (especially the end-capping) on prlonged storage. I would store the column in 100% acetonitrile. This is completely compatible with your mobile phase, so you can just run your method direct from the storage solvent.
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry