I'm pretty new to the forums and to liquid chromatography in general. I'm running a LCMS method on a C18 column and I was wondering if it is ok to wash and store columns in mobile phase containing acid, in this case 0.1% formic acid.

Thanks!

Posted: Wed Jan 16, 2008 3:48 pm
by goxy43
generally yes. just make sure that there is no salt on the column and that some amount of organic is in the mobile phase you use for storage. 10% acetonitrile with 0.1% FA will do the job.
success

Posted: Wed Jan 16, 2008 7:26 pm
by Victor
I would not store the column in an acidic solution. The pH of 0.1% formic acid is about 2.7. This can cause slow hydrolysis of the bonded ligands (especially the end-capping) on prlonged storage. I would store the column in 100% acetonitrile. This is completely compatible with your mobile phase, so you can just run your method direct from the storage solvent.

You can find your answer here.

Posted: Thu Jan 17, 2008 10:29 am
by Jackus