My questions still remain about using molecular sieve. Could it introduce contaminants into the mobile phase? And for to keep solvents dry they need to stand over the drying agents, which is practically impossible in HPLC since particulates in the mobile phase are not a great idea. I guess there are no simple answers to these questions.
Sorry - I should have been more explicit. In my experience, using molecular sieves with a subsequent 0.2 um filter works fine for 5 um particle size columns. I almost always use guard columns - unless client specifies otherwise - and didn't notice quicker pressure increases.
I used to just put about 100g of the molecular sieves beads into the 2.5L solvent bottle and leave for a week or so with occasional very gentle swirling, filtering as required. Molecular sieves are very friable, so care is needed.
If you want to dry the solvent quickly, it would become more of an issue, as you need to purchase ( or grind, then rinse fines away ) the MS as a powder ( eg GC packed column material ), and then ensure all fines are removed.
I recycled some MS used on volatile solvents ( dried under flowing nitrogen at 265C for 24 hours - after removing the solvent by vacuum ), but only used the regenerated MS on the same solvents.
I've never had a problem with contamination from MS, and have assumed that the method of manufacture would preclude most organic contaminants. For solvents, I used laboratory grade molecular sieves, I would be a little more wary of the technical/refrigeration grades, but they should also be OK.
Please keep having fun,
Bruce Hamilton