Victor:
You seem to be using the values of the measurements in a) and b) interchangeably, as if they are the same thing. Is this correct?
I
am assuming that extra-column volume and extra-column contribution to band spreading are the same thing. This is a
gross oversimplification, which is why I hedged my statements:
In practice, it's more complicated than that because unswept volume (like a poorly assembled fitting) is worse than a straight run of tubing, very narrow tubing is more prone to laminar-flow problems, and so on.
As a matter of practice, in all the years that I worked with the DryLab program (in which you can specify an "extra-column volume" which is actually treated as a band-broadening contribution), I found that "fudging" the value to make the predicted and observed chromatograms match in terms of peak width required a value which was in reasonable agreement with the estimated extra-column volume of the system.
Measuring the extra-column volume directly is appealing in principle, but runs into practical problems on a standard analytical scale system. At 1 mL/min, 16 microliters implies a transit time of less than 1 second, so that injection valve latency and detector response times become an issue. The y-intercept approach is more tedious, ultimately more reliable, and gives a measure of what we're really interested in, which is band-spreading.
That said, I'll reiterate my earlier statement that as a matter of practice, actually measuring the extra-column contribution to band spreading is not as useful as recognizing that we have to keep the extra-column volume to a minimum. This is especially true when we try to run very small columns on a standard system.