Dionex LC25 chromatography peak problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
HI guys,

I am experiencing a problem with Dionex LC 25 system for peaks. After loading a sample through an autosampler, the baseline is not normal and straight rather gives ripples throught the run and the peaks as well is not in baseline main baseline format but also surrounded by ripples and i am not able to get the retention time for any of these peaks. Guys Please help me out.
Thank you.

Imran
Is your mobile phase contaminated?
Have you got air in the mobile phase? (is it properly degassed?)
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth

did you check the depth of injection needle? any air aspirated? any problems with check valves of the pump? Try to run a "blank" sample w/o injection and see if the base line is still showing ripples.
goxy
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry