Peak area variations
Posted: Mon Jan 14, 2008 2:27 pm
Dear Members,
I experience peak area variations I cannot explain. It is RP chromatography, sodium phosphate monobasic/phosphoric acid buffer (pH=3): acetonitrile (70:30), isocratic for main peaks with gradient later to wash out garbage. SST is always excellent. Samples are bracketed by standards, pattern of injections -- std, two samples (first preparation), two samples (duplicate), std. Two injections from first preparation have the same area between them, two injections from duplicate preparation have the same area between them, but duplicate preparation samples always have smaller area than first preparation (RSD 3 to 10). Sample preparation is simple and no mistakes are made. I even tried to fill those vials (first and second preparations) with a sample from one preparation only -- the result is the same -- different areas. Equilibration of the column seems to be OK, new column makes no difference. The system seems to be OK. Can you help, please?
Al Bragi
Canada
Al Bragi
Canada