Ion Chromatography-High backgr. conductivity in Dionex DX500
Posted: Mon Sep 20, 2004 5:58 pm
We recently followed the cleanup procedure suggested by the Dionex manual, and we passed through our CS12 (configured with a CG12) solutions of 1M HCl (for 60') and 0,01M HCl (for 15' before and after the 1M solution). After that, the system was left to equilibrate with eluent running through for 1 hour (the manual suggests 30 mins), and then the suppressor was again connected, the guard column was placed before the column and the system was turnt to work again.
While the peaks look almost ideal, and separation seems to take place as it should (checking standards' labelled concentrations are verified with a discrepancy within 5%), the background conductivity remains as high as 6.0us (while it should normally be less than 2.0us), with suppressor working at 100mA. At the same time, the H20 dip at the beginning of the chromatogram has grown since earlier.
What would cause that high background conductivity? Could it be any leftovers of HCl as we suspect - but if it's this how can it be faced? Could it be any HCl leftovers that passed through the suppressor and contaminated it?
Any ideas/suggestions would be highly appreciated.