HILIC is an alternative for triflic and benzenesulphonic acid. A good thing with HILIC compared to ion exchange is the possibility of using high organic solvent/low ionic strength mobile phases which allows simple connection with MS. This will give you a higher sensitivity and also much less noise from column bleed.
To exemplify this, I've attached an application note on separation of organophosphonate nerve agent metabolites, that is phosphonic acids with different substituents. In this example (isocratic) you will see that as the analyte becomes more polar (less hydrophobic substituents) it gets more retained.
This application can easily be improved by a gradient but it is suitable as an illustration on how a small change in a side change can increase/decrease retention on the column.
http://www.sequant.com/sn/ufiles/SeQuan ... 00-18A.pdf
If you have not yet evaluated HILIC, I'd suggest that you try our ZIC®-HILIC phase, and if an alternative detection mode is of interest, I'd suggest you combine it with ELSD.