A couple of quick questions...
(1) So if we're injecting samples in splitless mode - let's say 1 uL of a sample in DCM with an inlet T of 250, head pressure of 10 psi in an inlet liner of 493 uL effective volume - this should give a 399 uL vapour cloud which should be fine. Now if one wanted to inject 2 uL instead, the vapour cloud would be 797 uL, which would probably cause backflash.
But let's say you are injecting in split mode, with for example a 50:1 split ratio - does the ratio apply directly now so that a 2 uL injection would be fine and reproducible? I'm a little confused on how to know whether backflash should be an issue with split injections and if so, at which point (injection volume, conditions and split ratio) does it become an issue...?
(2) Technically, all gases we use on our GC-FID are UHP grade. But the last few cylinders of air that were used were actual industrial grade. The thing is we haven't noticed any major problems with using this lower (and much cheaper!) grade... Is it possible that it only raises the baseline? By how much? Anyone have any experience with this?
Thanks,