I am using good Ultrahydrogel column 250, 1000 and 2000 in a series.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
I have run high MW PEO up to 13 million, what injection volume and concentration do you use? I find that we need to lower the concentration to less than 0.25 mg/ml and dissolve for 24-48 hours and then inject 50 - 100 micro liters.
We are a using cannon fenske solution viscosity not a viscometer detector for GPC. For GPC we are using a RI detector. I am definitely interested in more details, i would greatly appreciate any help you can offer. Thanks.When you say viscosity results are you referring to a batch mode viscosity measurement, like a cannon fenske solution viscosity, or a viscometer detector for GPC? I have found that the concentration of the solution viscosity measurement is usually higher than for a GPC experiment. Our High Mw PEO samples were run using GPC with very reproducible results using light scattering detection. If you need more details, please let me know.
What range of standards did you use for you calibration curve. We are currently using a narrow range of standards ranging from 500,000 to 1,250,000M MW. If I use this curve, I would have to extrapolate when analysing higher MW, so I am wondering if at high MW would the calibration curve still be valid.Our work was done on a Polymer Standards Service Suprema Aqueous GPC column with a 30,000 angstrom pore, 10 micron particle, 8 x 300 mm column. Flow rate was 0.3 ml/min, injection volume was 50 microliters, concentration and preparation of sample was as in my previous post. We found that if you tried to run the higher Mw PEO materials at flow rates greater than 0.3 ml/min they tended to shift to lower molecular weights, i.e. they all appeard to be near 2 million. Slowing the flow rate allowed the bigger molecules sufficient time to diffuse though the pores. These big molecules need tome to relax and flow through the GPC column. Also, we used a large inlet fritte than standard at the column ends in order to avoid shearing at the inlet and outlet of the column. Results were consistent with ultracetrification results. Link for the column is here:
http://pssgpcshop.com/mm5/merchant.mvc? ... de=suprema
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.