GC: Varian GC3800
Injector: Varian 1177 split/splitless
Split Ratio: 8:1
Column: CP-PoraBOND q fused silica
Detector: FID
We are using this set up to measure ethylene concentration of approximately 10 PPM and below. The first testing and application used a small sampling pump to draw samples from the system to the sample inlet on the side of the GC. The pump was left on at all time and the sample was then returned to the system. The flow rate was approximately 2 LPM. This set up worked beautifully and gave very reproducible results and the accuracy needed.
Going forward it is desired to manually inject samples into the GC. I.E. we wish to draw samples into a syringe at a remote or semi remote location and inject the said samples into the GC. I have tried injecting 500 uL of gas through the septa, results looked promising but reproducibility stinks. I have tried upping the volume to 5 mL again results looked promising but the reproducibility stinks. I have tried using both of these syringes to also inject said samples (500 uL and 5 mL) into the sample inlet port on the side of the GC. These were all ideas presented to me via the guy that trained us, I have very little faith in some of the things he says. So does anyone have any suggestions on a method of manual sample injection that might provide reproducible results? How much volume should be injected, where should the volume be injected, at what rate should it be injected, when should the start of the run be triggered (before, during, or after manual injection begins. Any help would be greatly appreciated.
D