ammonium bicarbonate Vs ammonia buffer
Posted: Sat Jan 05, 2008 1:38 pm
Hi
Currently i have a gradient HPLC run which uses 10 mM Ammonium bicarbonate buffer adjusted to pH 10.0 with ammonium hydroxide.
MP A : the above buffer
MP B: ACN
When i run gradient
time A B curve
0 95 5
5 95 5 6
17 50 50 6
36 50 50 6
37 95 5 6
I see a negative baseline after 8 minutes. the baseline goes from 0 absorbance units to -0.01. and then goes back to 0 when the gradient reaches 95 :5.
When i use 10 mM Ammonia (PH is 10.6) as my MP A. I don't see this behavior . But my late eluters specificity interchanged.
Column used is xterra, RP8 3.5 u, 4.6x150mm
wavelength 235 nm.
at 225 and below negative baseline is worse.
My question is why i see negative baseline with ammonium bicarbonate and why not with ammonia at 235 nm wave length.
Buffer capacity of 10 mM ammonia (pH10.6) is good or not. I am wondering to get the same specificity as ammonium bicarbonate can i prepare 5 mM ammonia ( does it have buffer capacity or not).
Please explain
MP B: ACN
When i run gradient
time A B curve
0 95 5
5 95 5 6
17 50 50 6
36 50 50 6
37 95 5 6
wavelength 235 nm.
Please explain