One reason for testing the column using the manufacturer's test mixture is to ascertain how your system compares to the test system.
It smells of instrument qualification – more than a column performance test. I would much rather handle system and column issues separately.
Maybe. But how many of your column suppliers detail their test systems on their certificates?. Few of mine do. For me, it's nothing to do with instrument qualification, it's about ensuring I understand how a brand new column performs on my system, and how it degrades during use on my system. IQ merely measures how my HPLC matches the manufactuers' specification, and should be designed to be column independent.
If you are a bulk buyer of columns, and a QC/QA user, your system suitability may detect an inadequate column, but it may not...
If my SST does not detect any insufficiency, then I’m fine! Why would I worry as whether or not the column would perform better or worse with regard to some compounds, I’m not planning to analyse?
The role of the SST is precisely to document an adequate performance of the system (including the particular column) with regard to the tested samples/compounds.
I noted that system suitability may detect an OOS column, but for users like me, where I use a column for a diverse range of analyses, the vast majority of which don't require a system suitability test, the manufacturers' tests are simple, effective, and relevant. YMMV.
I agree it's not that common, but I recall at least two situations where column performance did not match the certificate, and the failed column was exchanged. The replacement columns worked fine.
There you go: As you mention you recall a couple of situations where you found an inconsistency (I’m not sure whether or not it would have been an issue for your analytical method but let’s assume it would have)
Maybe the vendor would have exchanged these 2 columns anyway – assuming they didn’t perform as expected according to your SST.
The reason I instituted the testing of columns is because of an earlier situation, where a manufactuer refused to accept that a column was faulty because we had put samples through it before testing it. I had to swallow hard buy another and complete the project. I still try avoid columns from that manufacturer.
But most importantly you should compare the time invested (hence the cost) in testing every column with different eluents and compounds to the cost of a couple of columns you potentially would dispose of, every 5 years, or something like that. What is then most cost effective?
As I noted earlier, the vast majority of performance tests are simple isocratic systems, and easy to perform. The test solutions are stable, and the analysis typically only take about 15-20 mins. Some manufacturers, such as Phenomenex, may use their recommended storage solvent composition ( 65/35 CH3CN:H2O ) as the test mobile phase, so it's easy to do on most of their reverse phase columns. Other ratios of CH3CN:H2O are easy on many systems.
I only perform the test when I get the column, and when I suspect something is wrong. I also use it when a column forms a void and I repack it, often plates plummet to 20-30% of new, and come back to 90+%, which is fine for most of my work. Takes 5-10 minutes to fill a void in a column and compress it.
As I noted earlier, a column typically costs several hundred to several thousand dollars, and whether column testing time is justified depends on your environment. It is for me, but YMMV.
Please keep having fun,
Bruce Hamilton