I got a ghost peak in my chromatogram.
The gradient is:
Time (min) % Water % ACN Flow (mL/min)
0 50 50 1
11 0 100 1
25 0 100 1
30 50 50 1
35 50 50 1
It has fixed retention time of 28 min, and very nice peak shape. Its peak area (more than 4000) is independent on sample injection volume, sample concentration and sample solvent species. Its peak area is dependent on the purity of my acetonitrile mobile phase.
(HPLC grade, Peak area 4000, gradient grade, Peak area 400)
What are ghost peak origins?
1. Sample: Injecting different samples, different solvents should get rid of it. So NO.
2. pH electrode : NO, I did not adjust pH.
3. Water: NO, I used different water
4. Mobile phase?: should retention time be random? ACN mobile phase purity does affect its peak area. (most likely)
5. Column?: Peak area should change with washing, and its not observed that peak area is dependent on washing. I only have one column and it is hard to tell.
6. Injector?: by pass column, it should be there. I tried using a connector and it is there, but the area decreases. I tried 0 mL injection and it is still there.