[Question] Non-linear standard (calibrator) curves
Posted: Thu Dec 27, 2007 9:14 pm
Merry Christmas!
I have a problem that perhaps some of you might help me with
I am setting up a LC-MS/MS analysis of various hormones, but somehow the standard curves (12 points) are not quite linear. Here are the key points:
- I try to quantify over a pretty large range (1 - 2000 nmol/L).
- The highest concentration seem to fall a little (15-30%) too low.
- I have tried preparing the standards in H2O, MeOH, MeOH:H2O, bovine serum albumin and stripped serum - but the non-linearity seems consistent.
- My internal standards are deutered isotopes of the analytes and coelute exactly with my substances.
- ISs are in a rather low concentration (10-50nmol/L), as for most of my samples this will give approx 1:1 ratio between the substances and IS. This is a compromise as this method is not all about measuring "normal" human beings, but rather also quantify very high pathological values.
- I am doing fully automated cleanup online coupled with an analytical column (SunFire).
So...why have I ran into this kind of trouble? And how do I solve this??? Eh...
Perhaps I was to optimistic to expect perfect linearity over such a large range? Maybe it has something to do with the IS has way lower concentration than the hormones? I think (but have not fully tested this) that the situation improves by injecting lower sample volumes when running "high samples". Or maybe this should make me really reconsider my method, perhaps something is completely wrong?
Anyway...here is my suggestions of different workarounds. Which is the most appropriate?
- Rerun every sample beyond a certain concentration, but with lower sample volume.
- Rerun every sample beyond a certain concentration, but first dilute.
- Rerun every sample beyond a certain concentration, but add more IS to the sample and standard curve.
- Add more IS to every sample
- Apply a "quadratic regression" - this way it all seems to fit nicely. But I don't know the mathematics. Is this ok?
If would be nice if anyone here can point me in the right direction? Or better - have a definite answer!
Yours,
Anna
I have a problem that perhaps some of you might help me with
I am setting up a LC-MS/MS analysis of various hormones, but somehow the standard curves (12 points) are not quite linear. Here are the key points:
- I try to quantify over a pretty large range (1 - 2000 nmol/L).
- The highest concentration seem to fall a little (15-30%) too low.
- I have tried preparing the standards in H2O, MeOH, MeOH:H2O, bovine serum albumin and stripped serum - but the non-linearity seems consistent.
- My internal standards are deutered isotopes of the analytes and coelute exactly with my substances.
- ISs are in a rather low concentration (10-50nmol/L), as for most of my samples this will give approx 1:1 ratio between the substances and IS. This is a compromise as this method is not all about measuring "normal" human beings, but rather also quantify very high pathological values.
- I am doing fully automated cleanup online coupled with an analytical column (SunFire).
So...why have I ran into this kind of trouble? And how do I solve this??? Eh...
Perhaps I was to optimistic to expect perfect linearity over such a large range? Maybe it has something to do with the IS has way lower concentration than the hormones? I think (but have not fully tested this) that the situation improves by injecting lower sample volumes when running "high samples". Or maybe this should make me really reconsider my method, perhaps something is completely wrong?
Anyway...here is my suggestions of different workarounds. Which is the most appropriate?
- Rerun every sample beyond a certain concentration, but with lower sample volume.
- Rerun every sample beyond a certain concentration, but first dilute.
- Rerun every sample beyond a certain concentration, but add more IS to the sample and standard curve.
- Add more IS to every sample
- Apply a "quadratic regression" - this way it all seems to fit nicely. But I don't know the mathematics. Is this ok?
If would be nice if anyone here can point me in the right direction? Or better - have a definite answer!
Yours,
Anna