[quote]The Sigma-Aldrich product manager indicated that he often sees large, elevated baselines with inexperienced biodiesel producers. This can be attributed to two sources:
1. Incomplete derivatization of the sample with MSTFA. This can be caused by:
a. using the wrong MSTFA reagent;
b. not waiting for the reaction to complete;
c. waiting too long after preparing the samples;
d. excess water or alcohol in the sample that uses up the MSTFA.
Elevated baseline due to incomplete derivatization can be overcome by baking out the column.
2. Poorly made biodiesel that contains large amounts of unreacted vegetable oil. This problem cannot be overcome and the column and precolumn must be replaced.
He also provided the following recommendations:
Calibration standards and samples should all be prepped at the same time, using the same lot of derivatization reagents.
The MSTFA derivatization reagent should be purchased in 1-ml ampoules, and fresh reagent should be used every time you prepare standards and samples. The only MSTFA customers should use is "Sigma-Aldrich Derivatization Gradeâ€