Increased analyte response with liner design
Posted: Wed Dec 19, 2007 11:09 am
Hello,
I am separating a C8-C20 series of n-alkanes dissolved in chloroform on a narrow-bore 25m x 0.15mm i.d. x 0.25µm film BP-1 column with detection by TOF MS. 1 uL is injected at 1/10 split (1 mL/min He column flow, 40cm/s average velocity, 41 psi head pressure at start column temperature of 40°C) using a CIS-4 (Gerstel) injection port in constant temperature mode at 320°C. The liners used in this injection port have a nominal volume of ~200 µL. Under the injection conditions used the solvent expansion volume (for chloroform) will be ~160µL.
If I use a deactivated glass wool liner, I get 2-3X the peak height for all peaks (so no evidence for BP discrimination) compared to using a baffled liner. Why should this be, since both liners have the same volume? The outcome suggests I am losing sample out the top of the baffled liner - but this should not be so as liner volume > expansion folume and there is effectively a 10mL/min He flow through the liner. Is glass wool known to slow down the vapour cloud expansion process?
I've also tried a cold split injection with both liners where the injector temperature is ramped from 10°C to 320°C over 25s. This should reduce any liner overfilling effect that might happen during a hot split injection. However, with either liner, the response for all peaks is lower with cold split. The peak shapes for the more volatile alkanes are broader with the cold split technique, which is understandable given that they won't be efficiently trapped on the liner at 10°C. However, for C20 alkane, which should be fully retained on the liner at 10°C, the response is lower for either liner in cold split compared to hot split, even though the peak shapes are identical. I can't explain this either!
Any insight would be greatly valued.
thanks
Tony
I am separating a C8-C20 series of n-alkanes dissolved in chloroform on a narrow-bore 25m x 0.15mm i.d. x 0.25µm film BP-1 column with detection by TOF MS. 1 uL is injected at 1/10 split (1 mL/min He column flow, 40cm/s average velocity, 41 psi head pressure at start column temperature of 40°C) using a CIS-4 (Gerstel) injection port in constant temperature mode at 320°C. The liners used in this injection port have a nominal volume of ~200 µL. Under the injection conditions used the solvent expansion volume (for chloroform) will be ~160µL.
If I use a deactivated glass wool liner, I get 2-3X the peak height for all peaks (so no evidence for BP discrimination) compared to using a baffled liner. Why should this be, since both liners have the same volume? The outcome suggests I am losing sample out the top of the baffled liner - but this should not be so as liner volume > expansion folume and there is effectively a 10mL/min He flow through the liner. Is glass wool known to slow down the vapour cloud expansion process?
I've also tried a cold split injection with both liners where the injector temperature is ramped from 10°C to 320°C over 25s. This should reduce any liner overfilling effect that might happen during a hot split injection. However, with either liner, the response for all peaks is lower with cold split. The peak shapes for the more volatile alkanes are broader with the cold split technique, which is understandable given that they won't be efficiently trapped on the liner at 10°C. However, for C20 alkane, which should be fully retained on the liner at 10°C, the response is lower for either liner in cold split compared to hot split, even though the peak shapes are identical. I can't explain this either!
Any insight would be greatly valued.
thanks
Tony