ergosterol HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am attempting to quantify ergosterol using HPLC. The standards I am running produce a peak with a shoulder. is this a problem

If your standards have a shoulder, in a word, yes, it is a problem. First off, what is the purity of your standard? If it is supposed to be 100% pure, then you have a dirty column I would guess. If it is significantly less than 100%, I would suspect you have a co-elution issue and need to continue development to separate the second peak.

You could have other problems, but this is I would start if I were you.

Ergosterol might be difficult to dissolve in aqueous mobile phases, and you may have dissolved it in 100% organic. This can create peak distortion in aqueous mobile phases if you inject a large sample volume.

To see if this is the problem, inject less sample.
3 posts Page 1 of 1

Who is online

In total there are 329 users online :: 1 registered, 0 hidden and 328 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 328 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry