Headspace linearity issue
Posted: Tue Dec 18, 2007 4:56 pm
I am working with a HT250D autosampler, which has a heated syringe, attached to a 6809N. I am trying to develop a PQ method for linearity.
I've been using a 5m x 0.1 mm i.d. uncoated column since this type of column has been used by engineers who've performed previous OQ's. My sample is 10 ppm ethanol in water where 2 mL of the sample is in a 10 mL headspace vial. The injection range is 0.4, 0.8, 1.2, 1.6, and 2.0mL. One of my better runs gives an R sq. value of 0.966 for the 5 points. For the first 4 points, linearity is 0.997.
Area (pA*s)
4539.04004
6119.06641
7547.75293
8765.08496
9183.38965
I'm getting close but not close enough. I've change liners, flows, splits, injections speed and haven't improved. I've increased the flow (0.7mL/min) to increase the inlet pressure to get a more compact sample plug. I decreased the injection speed to 8 mL/min keeping it to a little less than half of the total inlet flow.
Headspace: Oven 70C, Syringe 75C, Injection speed 75 mL/min.
heating time 15 minutes
GC:
Straight 4 mm i.d. liner (Agilent 210-3003-5)
Inlet temp 150C
Split 30:1
Inlet pressure 30.46 psi
Total Flow 21.4 mL/min.
Column flow 0.6 mL/min.
Oven 40C for 2min., ramp to 150C to burn off any water vapor in the column
I can provide other conditions if needed. Any input is greaty appreciated.
I've been using a 5m x 0.1 mm i.d. uncoated column since this type of column has been used by engineers who've performed previous OQ's. My sample is 10 ppm ethanol in water where 2 mL of the sample is in a 10 mL headspace vial. The injection range is 0.4, 0.8, 1.2, 1.6, and 2.0mL. One of my better runs gives an R sq. value of 0.966 for the 5 points. For the first 4 points, linearity is 0.997.
Area (pA*s)
4539.04004
6119.06641
7547.75293
8765.08496
9183.38965
I'm getting close but not close enough. I've change liners, flows, splits, injections speed and haven't improved. I've increased the flow (0.7mL/min) to increase the inlet pressure to get a more compact sample plug. I decreased the injection speed to 8 mL/min keeping it to a little less than half of the total inlet flow.
Headspace: Oven 70C, Syringe 75C, Injection speed 75 mL/min.
heating time 15 minutes
GC:
Straight 4 mm i.d. liner (Agilent 210-3003-5)
Inlet temp 150C
Split 30:1
Inlet pressure 30.46 psi
Total Flow 21.4 mL/min.
Column flow 0.6 mL/min.
Oven 40C for 2min., ramp to 150C to burn off any water vapor in the column
I can provide other conditions if needed. Any input is greaty appreciated.