Baseline rise with T-programming
Posted: Mon Dec 17, 2007 9:57 pm
So I can't seem to find a clear answer about this, but from searches I gather that baseline drift with temperature-rise in GC-MS is normal and attributed to column bleed? Can anyone clarify this for me? I mean, the column's upper temperature is 300 degC, why would it bleed so? What are the mechanisms that cause this to occur...?
And how much is actually "normal"? I've put up a screenshot of the chromatogram (TIC) we collected by injecting 1 uL of methanol [inlet = 250; split injection 60:1; flow 1.2 mL/min; T-program = 50 for 1 minute, then ramp (20 degC/min) to 100 (hold 0 mins) and ramp to 275 (7.5 degC/min) and hold 5 mins with a post-run of 2 mins at 300 degC.]
Screenshot can be seen here :
http://aix1.uottawa.ca/~rclemen2/MeOHinj.jpg
Thanks!
Roxanne.
And how much is actually "normal"? I've put up a screenshot of the chromatogram (TIC) we collected by injecting 1 uL of methanol [inlet = 250; split injection 60:1; flow 1.2 mL/min; T-program = 50 for 1 minute, then ramp (20 degC/min) to 100 (hold 0 mins) and ramp to 275 (7.5 degC/min) and hold 5 mins with a post-run of 2 mins at 300 degC.]