also remember that if you have a problem and that an auditor finds that this is the cause, he will not go to your vendor, he will ask you why you decided to use a product from a vendor that did not match your needs for regulation. (Italics added)
Not all specification are meaningful to all people. Whether the column meets all the vendor specs is (from my point of view) irrelevant. What matters is whether it meets
my specs. If it doesn't, back it goes. And if I were so naive/stupid/desperate as to use it anyway, then an auditor would be quite justified in "busting" me.
If it
does meet my specs, then there's no problem.
In this specific case, my question would be whether or not the exact particle size range was written into lcguy1's specification.
- If it was not, and if the column lets him meet "system suitability" for his method, then I would say he was safe from a regulatory perspective.
- If it was not, and the column fails "system suitability", then the column is unsuitable (blinding flash of the obvious!

)
- If it was part of his spec, and the column fails system suitability, then the column is unsuitable (even more blinding flash of the obvious!

) .
- If it was part of his spec, but the column lets him meet system suitability, then he's in a gray area. At that point, what to do depends a bit on the regulatory agency:
USP basically says that you can use any "equivalent" column so long as you pass system suitability. Just to put the importance they attach to particle size, the "L1" category of equivalent columns is defined as "Octadecyl silane chemically bonded to porous silica or ceramic micro-particles, 3 to 10 μm in diameter." (USP 27, sect 621).
The US FDA (for internal use) says: "Particle Size (HPLC): May be reduced by as much as 50%." (ORA-LAB.5.4.5, Attachment A).
I would assume a column vendor does not have to operate under cGMP, but is most likely under ISO 9xxx. They may be in full compliance with their own procedures and SOPs by releasing the column with a disclaimer on the COA and letting the customers decide what to do.
[As a side note, I'm convinced that a manufacturer of life jackets could be ISO-9xxx compliant with a line of concrete-filled life vests, so long as they used a consistent grade of concrete and had documented paperwork that the next-of-kin could use in filing claims

]