Helium has a wider range of flow for superior chromatography than does nitrogen
[Hydrogen] has an even wider range of flow as a carrier gas for chromatography
Rod,
I agree that ABSOLUTE range of acceptable flows in THE SAME COLUM is the widest for hydrogen and the narrowest for nitrogen. However, there could be another take on the implications.
According to Golay equation for a column plate height, the loss in a column efficiency for any gas depends on the ratio of actual flow to optimal flow. In other words, the loss depends on RELATIVE (NOT ABSOLUTE) difference between actual flow and optimal flow. For example, if flow of any particular gas is twice as high compared to optimal flow of that gas then efficiency of the same column would be the same whether the gas is hydrogen, helium or nitrogen.
One can conclude that
1. It is the RELATIVE (NOT ABSOLUTE) range of acceptable flows that mostly affects a column optimization.
2. Because there are very few (if any) practical reasons to choose the relative acceptable range of flows differently for different gas, there are very few (if any) practical reasons to consider this factor at all in choosing the carrier gas.
Example.
Optimal flow in 0.25 mm column is estimated as
2.5 mL/min for hydrogen
2 mL/min for helium
0.63 mL/min for nitrogen
Suppose that, based on some set of considerations, helium flow of 1.5 mL/min to 3 mL/min is considered to be acceptable. This translates into the range of 1.9 mL/min to 3.8 mL/min for hydrogen, and 0.47 mL/min to 0.94 mL/min for nitrogen. Typically, it would be acceptable to choose the range of flows for each gas according to this list, and to scale the flow rate of each gas accordingly. As a result, the plate number in the column would be the same regardless of the gas.
All in all, it is reasonable to suggest that carrier gas should be chosen for its affordability, speed of analysis, safety, etc., but (as a rule) not for the range of its flow rates.
lmb