We just got a brand new metal capillary column from restek and we're getting ready to install it. However, we need to condition it. We have an FID and the manual has said to not connect the other end of the column to the detector when conditioning. Although we're using Helium as the carrier gas I am concerned about leaving the other end out in the oven or in the open air while it is conditioning, as when we are done we will to cap it, turn the unit off and uncap it anyway install it onto the detector anyway once it is cool.

Why do all the manuals say leave it unattached to the detector while conditioning?

The primary reason is the bleed off the column as it is first conditioned will contaminate your FID and increase the noise level, raising the limit of detection of your detector.

Leaving the column in the oven is no more a problem than leaving the end of the column in the detector as it is flowing air and hydrogen to the flame tip.

Just make certain the column end does not enter the space occupied by the oven fan.

best wishes,

Rod

thanks, rod. now that i think about it i can see how that can happen.

one more question: before i put it into the detector should i leave the flow of helium on? or should i just cap it and install it as quickly as i can and then run helium through it again?

Leave the flow of helium on. Any 'capping' you do may contaminatethe column end and you wouldn't want to put a dirty end into the detector.

best wishes,

Rod

(1) Capillary columns most often are put through QC test which includes running a sample, and print-out of that is most often included, so the column has already been pre-run.

(2) My guess is that the part about not connecting to an FID is a hold-over from the old packed column days and of less importance with capillary columns (with smaller amounts of phase) and bonded phases.

thanks again for all t he info.

yeah we got the QA report from them and it looked golden on the baseline. i'm still a little concerned that i should condition it, b/c of the heat sealed ends. how far down should i score it, approximately? should i still condition it anyway?

I would cut off 3 to 5 inches as a rule. Rules are made to be broken. Rules are also made to be adjusted.

best wishes,

Rod
7 posts Page 1 of 1

Who is online

In total there are 116 users online :: 0 registered, 0 hidden and 116 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 116 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry