Hi all,

I have been trying to quantify a mixture of antibiotics in human plasma using UPLC-MS/MS. During method development, I am always left with excessive (somewhat crystalline) contamination of the ESI-chamber and cone. I have tried rinsing all tubing with IPA/MeOH/AcN/H2O and preparing fresh mobile phase before each run. I have avoided the use of non-volatile buffers and have diluted plasma to minimize the level of endogenous salts. What do you think: is it my sample prep? Should I go over to SPE or should this protein prec have worked? Are there any other causes I haven't thought of?

Sample prep:
Tenfold dilution of plasma with aqueous 0.1% FA in water, followed by precipitation with 10% TCA, containing IS (1 part dil plasma to 1 part 10%TCA). after that left for 20 min at 4C, followed by 5 min centrifugation at 14000rpm and transfer of supernatant to autosamplervial.
I've tried different protein precipitation methods, but TCA gives the best chromatograms; a higher conc is not an option since ion suppression starts to kick in.

LC: BEH-column with in-line filterunit (0.2um), flow=0.5mL/min, column temp=50C
Gradient: A = 0.1% FA in water, B = MeOH, 10% to 90% B over a period of appr. 2.5 min. Injection volume = 30 ul

Thanks in advance
Maurice

Hi Maurice,

we have some bad experiences with TCA in combination with MS. Acetonitrile precipitation of proteins can work well. But I would go for SPE to minimize the fauling of the source.

regards Bert

A plasma sample with just a protein precipitation will sooner or later clog your analytical column. I do not recommend this. I suggest to look into more sophisticated sample prep methods, such as SPE.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry