N2, O2 CO2 and H2 analysis with HID detector

Posted: Mon Nov 19, 2007 7:27 pm
by fisio625
We performed an analysis for N2, O2, CO2, H2 with carboxen 1010 plot column (25467) by using Agilent 6890 equipped with Helium Ionization Detector and Flame photometric Detector. We made a calibration with a standart gas. And then made a injection with samples (sample includes only N2, O2, CO2, and H2).
There is an area repeatability problem with standart. However there is no
area repeatability problem with samples. %RSD value is in the acceptable
ranges.
Our standart gas includes N2, O2, CO2, H2 and sulphur compunds (SO2, H2S,
Ethyl mercaptan, methyl mercaptan..vs.). There are 10 sulfur compounds in
the standart gas.
Could sulphur compounds effects the repeatability?. Because the samples has
only N2 O2 CO2 and H2 and no problem with them.
Best Regards

Posted: Mon Nov 19, 2007 8:30 pm
by chromatographer1
Yes, they can affect the analysis. The sulfur compounds are large in comparison to the fixed gases and they can occupy the pores in the 1010 column. They can also react with the oxygen and hydrogen in the standard or with the column surface given the right conditions.

best wishes,

Rod

Posted: Tue Nov 27, 2007 4:01 pm
by fisio625
thanks a lot