I have a problem with one of our laboratory's old CE GC's.
Without any warning chromatograms from "Carlo Erba (then Fisons/then Thermo) GC 6000 vega 2 series" went "blank".
Chromatograms from routine analysis and from blank/standard samples happen to have very small peaks or no peaks whatsoever.
I have checked column's for leakage (DB-624 and DB-XLB), septa, injector, methanol-whashed detector and still have no clue, why did those peaks disapear. There are small peaks from solvent and toluen standard, from time to time, but their area is 100x smaller than it should be.
What could trigger this behavior?
- A leakage in oxygen/air line (we had that occasionally)?
- Hardware problems with detector or electrometer? (Fissons/CE EL 980)
- Bad split/splittless ratio setup? (there was no change in settings - may only be some leak in a support line)
- Any other source of the problem? (except the usuall PEBKAC-like answer
How to solve this? Our service vendor doesn't really want to bother with our "much too old" GC.
Thanks in advance,
Dan
PS,
System & settings are:
CE GC 6000 vega 2 series + FID detector + Electrometer Fisons El 980 + integrator CE Instruments DP 700
mode: split/splitless (40sec after injection);
solvent - CS2, standard - toluen; usual analytes - volatile compounds
inj.temp: 150'C
det. temp: 250'C
carrier: He, 30kPa
air pressure: 100kPa
H2 pressure: 50kPa
column: DB-624, ID 0.53mm, length:30m, c.flow=20mL/min (back)
oven program:
40'C - 1 min plateau
1st rise: up to 100'C with 5'C/min - 1min plateu
2nd rise: up to 230'C with 10'C/min - 1min plateau