criteria for the stock solution stability?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,

I am working on RP-HPLC method for a drug substance. The main peak area of the stock solution in MeOH was found stable in 24 hours. The assay result of the method showed good linearity, precision and accuracy. However, while I checked the impurity profile, a small decomposition peak appeared in the stock solution (probably produced by reaction with MeOH) and inceased to nearly 1% of the main peak area in 24 hour. Should I exclude MeOH as solvent for preparing stock solution? Moreover, could I still use MeOH for organic phase of MP?

Tons of thanks!

(1) Make up standard solution every day
(2) Evaluate whether refrigeration of the standard solution helps
(3) Since you're still in the experimentation phase would an alternative solvent like acetonitrile help, hurt, or no change?
(4) If you change to a different sample and standard solvent, could you still use methanol in the mobile phase - I'd say YES, as the analyte would only be in contact with methanol (in the column) for a few minutes, so any degradation would be miniscule
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry