Flow rate for packed column Supelpak-S

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I have a Supelpak-S (18' packed) x 1/8" OD fitted in a HP 7890A. I want to condition the column, but I noticed that my pre-column rate is 30 mL/min and my post-column flow rate is 10 mL/min. Will conditioning the column fix this - and is 10 mL/min an OK flow rate if I bake it overnight at 150oC?

I have another spare packed column - Chromasil 330, not yet preconditioned and I notice the same drop in flow rate. My guess is that this decrease is a feature of non-conditioned columns, but I thought it was worth to check this, before conditioning.

All answers appreciated.

You should call Supelco technical service. You will get the correct answer and the call is free.

800-359-3041

My guess is that you have a serious leak in your instrument. If you do then conditioning the column at 150°C can damage it. DON'T DO IT. Give Supelco a call first.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 127 users online :: 0 registered, 0 hidden and 127 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 127 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry