There is one problem. On one of the instruments, when DMF is the diluent, we see breakdown products (probably dimethylamine or methylamine)....a peak near methanol that varies in size and has considerable tailing on 624 type columns.
I've changed liners, inlet seals, septa, but nothing seems to help. Would an active site somewhere in the G1888 be the source of this breakdown? It doesn't matter what temperature any of the system's components is, I still get this peak...and only on this system. Adding 1 ml 0.1N HCl to the HS vials eliminates my amine problem, but I can't always do this.
If it is the headspace sampler, what can I do to eliminate the active site. I'm assuming it's the needle, but can it be in the valve or transfer line?
Burt