Cyno column mechanism

Basic questions from students; resources for projects and reports.

5 posts Page 1 of 1
Hi

Could any one tell me what is the underlying mechanism behind retention in “cynoâ€

I'm not sure about "cyno" columns, but cyano columns have that -C=N: sticking out, which is a fairly polar moity, so they are considered a normal phase column.

(OK, so underscoring an equals sign doesn't really look like a triple bond....)
All standard disclaimers apply: This post reflects personal opinion only and not the policies of my employer.

I assume you mean indeed cyano column.
this column can either be used in normalphase mode or reversed phase mode. Compared to C18 phase the cyano phase has a different selectivity. We used it for compounds with high pi electrons content.(aromatics).

philippe

Hi

If it can behave as both normal and reverse phase..? in which way I go ahead with method development.??
Could any one tell me the underlying principle of separation…? Philippem mentioned high pi electrons content..may I know how does this factor contribute in separation mechanism???

"Normal phase" and "reversed phase" refer to the relative polarities of the column and the solvent. If the column is polar and the solvent is non-polar, it's "normal-phase" chromatography. If the column is non-polar and the solvent is polar, it's "reversed-phase" chromatography.

An archetypical normal-phase column would be something like "bare" silica gel. An archetypical reversed-phase column would be something like octadecyl (C18) groups bonded to a silica support.

Most "cyano" columns consist of something like cyanopropyl groups bonded to a silica support, That cyano group has intermediate polarity, so if you use it with something like hexane/methylene chloride mobile phase, you will be doing normal-phase chromatography. If you use it with water/methanol mobile phse, you will be doing reversed-phase chromatography. (I would not recommend using the same physical column because of problems changing over between immiscible solvents).

phillipem was referring primarily to the use of cyano columns in reversed-phase, where the selectivity is somewhat different from that of C18 columns due to secondary interactions with pi and non-bonding electrons. The fundamental retention mechanism is hydrophobicity, just as in any other reversed-phase separation; the secondary interactions "tweak" retention a bit to move peaks a little further apart (or closer together!).

The mechanism in normal-phase is somewhat hazier (in my mind, at least); the simplest explanation is dipole-dipole interactions with analyte molecules.

Of course, all of the above explanations are over-simplified :wink:

If you can, get a copy of "Practical HPLC Method Development" by Snyder, Glajch, and Kirkland ( http://www.lcresources.com/resources/resbooks.html ). That has much more thorough explanations of what's going on.

The bottom line: you first have to decide what kind of chromatography you want to do (reverse- or normal-phase), then select the column, not the other way around
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry