Solvent for sample injection in LC-MS/MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Hi,
Is it possible to use 0.01M hydrochloric acid as final solvent for sample injection on a ESI LC-MS/MS?
Thanks

The pH of 0.01M HCl is 2. That is about the lower limit that most silica based columns can handle. Polymeric type columns can withstand a lower pH. Since you only inject a small volume (typically 1-50uL), it should not cause much problem with your column or the MS. It might decrease the lifetime of your column slightly, but not much.

Sassman,
Thank you for your answer.
There was one good article published in Elsevier "Importance of injection solution compositin for LC-MS/MS methods" (Journal of Pharmaceutical and Biomedical Analysis 26(2001) 753-767). You may refer this. If you could not get it, provide your mail address, I will mail you when I see your post.
Dear Jag,
I tried to open the link which you had sent related to the injection solvent in MS. Kindly send the paper to my following mail id: jbshan@gmail.com

thanx
Shan
5 posts Page 1 of 1

Who is online

In total there are 218 users online :: 1 registered, 0 hidden and 217 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Google [Bot] and 217 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry