The average plate count for Primesep columns is over 80K plates/meter. Methods in the link are less then 6 minutes with full base line resolution. For sharper peaks we suggest to use a longer column and play with pH and buffer concentration. For amines with double charge (histamine) pH and buffer is more critical as it interacts more strongly with the stationary phase.
With mixed mode chromatography you can achieve extremely sharp peaks by playing with pH. If you can "afford" gradient you will observe GC type peaks with double/triple gradients (pH buffer, concentration and ACN) due to focusing effect. This is not possible with RP or ion-exchange
http://www.sielc.com/application_122.html
"Plate count" (which is not an accurate definition in this case) is over 1.5M plates/meter. It would be interesting to see histamine on Dionex column (peak shape retention control, etc.)
P.S. There are many tricks to present chromatograms, if I create 40 minute scale for the one I cited above peak shape will be much much sharper, but this is a fast method to save time. Time is …very precious.