signal to noise ratio?

Posted: Mon Oct 29, 2007 6:16 pm
by zimanli
HOw to get signal to noise ratio and resolution between two peaks using Chemstation for Agilent 1100?

thanks

Posted: Mon Oct 29, 2007 9:32 pm
by zokitano
zimanli,

When you're creating report in Data Analysis go to the menu bar Report/Specify Report and then choose the type of the report: Performance + Noise (if you want to have the noise in your report with resolution etc)

Hope this helps

Posted: Tue Oct 30, 2007 2:07 am
by zimanli
thanks
have a nice day

Posted: Tue Oct 30, 2007 1:06 pm
by zimanli
I have signal/noise column, but it is all - without numbers.
DO I need to set any parameters to get signal to noise ratio for each peak?

thanks

Posted: Tue Oct 30, 2007 1:56 pm
by zokitano
Ok, I forgot to tell you to follow the entire procedure when you're creating a new method...

Anyway, in order to obtain your S/N values in the report you'll have to do the following:

1)In Method and Run Control click on the Menu bar Method

2)From the fall-down menu select Edit Entire Method

3)Input your method parameters (if you haven't done it already) for flow, injector, column compartment, signal...

4)When you'll get to the parameter: Specify Report in the window select the Report Style to be Performance+Noise and click OK to close the window

5)At the end of the method editing you'll come to a box that specifies the parameter Signal to Noise. Enter the time windows in which you want S/N to be measured by the software.

6)At the end save the method and run your analyses

7)If you want to change the time windows for S/N measuring during the optimization of your method just start the procedure from 1) and follow the instructions.

In your report you'll get the numbers for S/N for the specified time window(s) that you have already set in the method.

I hope this helps
Good luck