for this chromatogram
Image

which is for blank solvent (ethyl acetate) by Agilent GC-MS 5973B and HP-5ms (30m,25mic,0.25mic) for 1 min @ 60 C then 10 C/min for 300 C where kept for 3 min.
spliteless, 3.5min solvent dealy, 1 ml/min He flow
scan from 50 to 550


why there are no peaks?
why it's going up at the end?

Che313,

It is going up at the end because of column bleed. Look for 207 ion in scans out there.

Could be that the main ion is below 50 amu but even so, the secondary ions from the solvent should show up. No peak is probably because your ethyl acetate comes out in the solvent delay. Put a know amount of something heavier in there like dodecane at 50 micrograms/ml. Out of curiousity, what does the scan at 3.6 show?

Best regards.

Why do you expect to see peaks when you inject pure solvent ?
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry