I also think that in spectrophotometric methods such as UV, Fluorescene and IR , matrix effect or instrumental effect will always give signals (+ive or -ive) when no analyte is present.
Adding a non-experimental point 0,0 or "force to origin"
alters the real result from the experiment and make the cal. curve biased from the real experimental condition.
However, in some situation like chromatographic method , e.g. GC-FID
My thought is that :
two instrumental situation should be corrected before conducting GC-FID experiment:
1, no analyte but signal (or Peak) present at the retention time of the analyte
2, negative signal occur at the retention time of the analyte when no analyte presents
that means GC-FID system should be error (or abnormal situation) free (at least the errors of above) before doing experiment.
Then we use the above error-free system to calibrate curve using y=mx+c method.
However, if (1) the calibration curve generates x-value > zero (e.g. amount) when y-value (e.g. peak area) equals zero,
(2) the calibration curve generates negative x-value
Is it contradicting with the instrumental correction above ?
Adding point (0,0) to the curve will not solve the problem of -ive x-value , so only "force to origin" solves both -ive x-value problem and "zero analyte giving zero calculated amount" problem.
Does the "Force to origin " treatment make the cal. curve more closer to the real experimental condition ?