MPA: H2O:ACN: Formic Acid=99:1:0.2 (v/v/v)
MPB: ACN:Formic Acid= 100:0.2
Wavelength: 275 nm
Sunfire C18 column
Gradient
Time A B
0 100 0
8 100 0
25 60 40
27 60 40
28 60 40
40 100 0
Different detector has different baselines. On systems (2) with VWD UV detector, the baseline started hiking at 12 minutes. It makes sense since the gradient starts and ACN percentge increases. In contrast, on systems (2) with DAD or Multiwavelength detector, the baseline started going down at appox 12 minutes. All the analyte peaks did elute at the same time. Same column and mobile phase were used for all test injections. All four Systems are Agilent. Have you seen this before? What's behind this?
Next I will try on a system with VWD and DAD (or MWD) in series to confirm if it is really due to the detector.