"Olefin trap" and iodoalkanes, how well will it tr
Posted: Fri Oct 19, 2007 4:09 pm
by jonjon
Wondering if anyone has an idea of how well an olefin trap (silver impregnanted) will trap iodoalkanes (C1-C6) from an air sample collection. We are looking into some large volume air sampling (1000s m3) and I am looking into some prefractionation approaches prior to GC-MS analysis. Will likely collect initially on Carboxen type, then thermally desorb to secondary trap (olefin?), then to GC. Ideally we would like to inject the entire air sample collected.
Jon
Posted: Fri Oct 19, 2007 6:04 pm
by chromatographer1
Do you think the silver will react with the halogenated compounds?
I would test it and see. If it is reactive then you will have low recoveries.
I would try Carboxen 1016 for your trap.
best wishes,
Posted: Sun Oct 21, 2007 9:36 am
by MikeD
Rod,
Is there any water isotherm data available for Carboxen 1016 ?
Posted: Mon Oct 22, 2007 12:06 pm
by chromatographer1
I have no data that I can share with you. I can tell you that water is retained very little on Carboxen 1016 and should elute at the back side of the fixed gases (close to CO2).
I hope that is helpful.
best wishes,