Hi.

I am trying to learn the SPME method, and then I have to change septum on my GC-MS (HP 6890 Series II gas chromatograph equipped with a HP-5 MS 5 % phenyl-methyl-siloxane capillary column (30m x 0.25m x 0,25 µm film thickness) connected to a HP 5973 Series mass selective detector). When I replaced the old septum (Agilent 5183-4757) with the new recommended Thermagreen LB-2 septum for SPME, the "front inlet pressure shutdown"-message appeared.

I tried to restart to GC-MS twice, but the same happened.

Then I replaced the old septum, restarted the GC-MS again, and everything was fine. The inlet temp was 270 as usual.

Can someone help me with this problem? How can I get my Thermogreen septa to work?

BÃ¥rd

Best guess is that the tightness of the septum nut is critical - and you are probably screwing it down too hard.

Peter
Peter Apps

baardlassen,

I have had this problem on the 6890's and I believe Peter is right on the money. Probably too tight. With the pressure turned off (and the oven cool) pull the old septa, push the new one in and re-install the septum nut. Then turn on the pressure and start tightening the septum nut. When you see the pressure really start to jump, only 1/2 turn more. Has always worked well for me.

(and you don't have to clean pieces of smashed septum out of the inlet nut!!)

Best regards.
3 posts Page 1 of 1

Who is online

In total there are 249 users online :: 1 registered, 0 hidden and 248 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Amazon [Bot] and 248 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry