I am seeing two contaminat ions on LC-MS, 227.2 and 453.2. It's the ABI QTrap 2000, EMS positive scan mode. Mobile phase containg water, acetonitrile and formic acid, ammonium formate.

I suspect these ions are plasticizers, but I didn't filter the mobile phase. Any boby know waht these ions are and what could be the contamiant scouce?

Thank you.
I am not sure about 227.2, but 453.2 is a palsticizer. I also got 453.2 when I used formic acid packed in plastic bottle. It is known to be the adduct of diisooctyl phathalte with sodium and acetonitrile (Diisooctyl phthalate+Na+CH3CN).

If any one of reagent used for mobile phase is packed in plastic container then you may get this ion even if you don't filter the mobile phase.

Tahnk you, Jag.

I did use formic acid packed in plastic bottle. So that's the problem. I also used water purified through a water system, the 227.2 could come from the water.

Thank you for replying my post.
3 posts Page 1 of 1

Who is online

In total there are 209 users online :: 0 registered, 0 hidden and 209 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 209 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry