Posted: Mon Oct 15, 2007 6:57 pm
Preface: I am in the API pharmaceutical industry and am working on improving an LC method that is used for release testing.
I am substituting one L1 column for another L1, changing manufacturer as well as particle size from 5 to 3.5 um. No other method parameters are changing.
What validation elements would you include in your protocol? We have no SOP that covers this, although it sounds like a good one for me to write. Personally, I think all elements should be done (other than stability related elements), but management gets testy with instrument and Wanda time.
I will do the obvious - accuracy as injection reproducibility, linearity, specificity. should I determine LOD/LOQ? If so, then I obviously will do accuracy as spike recovery. What about robustness? I imagine interlaboratory precision needs to be evaluated since this method is used by Quality group.
I thought I would ask the LC gurus of the world their opinion so I can substantiate the required Wanda/instrument time.
thanks!
I am substituting one L1 column for another L1, changing manufacturer as well as particle size from 5 to 3.5 um. No other method parameters are changing.
What validation elements would you include in your protocol? We have no SOP that covers this, although it sounds like a good one for me to write. Personally, I think all elements should be done (other than stability related elements), but management gets testy with instrument and Wanda time.
I will do the obvious - accuracy as injection reproducibility, linearity, specificity. should I determine LOD/LOQ? If so, then I obviously will do accuracy as spike recovery. What about robustness? I imagine interlaboratory precision needs to be evaluated since this method is used by Quality group.
I thought I would ask the LC gurus of the world their opinion so I can substantiate the required Wanda/instrument time.
thanks!