We are using a Waters Alliance 2695 system with Waters 2487 dual wavelength detector. We also have a Waters 2996 PDA detector on the same system and have collected data with both; the "ripple" looks identical between the two, which says to us the detector is not the problem. We have to do related impurities analysis at a fairly low concentration and need the sensitivity of the 2487; the contortions of the baseline effectively erases the difference between the two for us in terms of their usefulness.
We have been back and forth with Waters on this and tried many experiments they have asked of us including changing the suppliers for the mobile phases used (we were already using HPLC grade solvents; we also tried substituting MilliQ for the water). The solvents are modified with TFA, we have been adding J. T. Baker HPLC grade TFA shipped in 1 mL glass ampoules. We have tried zero volume injections, injections with nio column, etc. The problem disappears in cases where we inject with no column, but using a different column doesn't make the problem go away.
The mobile phases in all cases:
- * Mobile A: Water + 0.05% Trifluoroacetic Acid
* Mobile B: Acetonitrile + 0.05% Trifluoroacetic Acid
Column (for the examples below): Phenomenex Luna C18(2), 100A, 250x4.6 mm, 3 um packing.
UV detection at 220 nm. 10 uL injection (although identical effect seen with 0 uL injection).
Method 1:
Code: Select all
Time (min) Flow (mL/min) %A %B
0 0.8 75.0 25.0
30.0 0.8 60.0 40.0
50.0 0.8 30.0 70.0
51.0 0.8 5.0 95.0
55.0 0.8 5.0 95.0
56.0 0.8 75.0 25.0
65.0 0.8 75.0 25.0

Method 2:
Code: Select all
Time (min) Flow (mL/min) %A %B
0 1.2 95.0 5.0
10.00 1.2 75.0 25.0
15.00 1.2 5.0 95.0
20.00 1.2 5.0 95.0
21.00 1.2 95.0 5.0
30.00 1.2 95.0 5.0

These are methods I am transferring from a different lab and I don't have any choice about any of these parameters.