in our lab we are running Epa 8260 method with purge e trap OI Eclipse plus 6890/5975 msd.
The problem is IS area response: the area go down during the batch and is very unstable.
When we have a calibration control spiked at 5ppb rise, while with blank or clean sample go down.
The IS are added automatically from Archon Autosampler with Valco valve at 1 ppb final concentration.
Our system has turbo pump, and the source is just replaced with clean one.
The column is DB-624, 30m, and we work in split mode with 20 ml of flow.
Does anyone has some suggestion?
Thank you
daniele
------------------