I am developing a LC-MS/MS (ESI) method for determining tetracycline in biological samples, but I am facing problems regarding poor reproducibility of analytes peak areas in standard solutions. Relative standard deviation of the internal standard (demeclocycline at a fixed concentration) was about 10% for 18 injections (3 replicates of six calibration point levels injected randomly), but it is now varying systematically arriving 35%. I am using a gradient elution on a 3µm Pursuit C18 analytical column (Varian), dimensions 100x2.0mm, with methanol, acetonitrile and water all containing 0.1% formic acid. The organic mobile phase reaches 70% in the gradient. Before the calibration curve, blank samples (solvent) are always injected in order to stabilize the system. Retention times are far from the dead volume and reproducible. Injection sample composition effect was evaluated and the injection sample composition is weaker than initial mobile phase (20% methanol). Other parameters were checked, such as vial and septum adsorption of analytes and autosampler problems (it was qualified, set at 4°C and methanol acidified with 1% formic acid is being used as rinsing solvent in a needle wash program). Carryover was not observed injecting blank samples after standard solutions at the same analytical conditions. Could the analytes be sticking to any parts of autosampler, to peek tubing or to the column? Why did it work before and does not work now? What kind of rinsing solvent do you suggest? Do you think I have to modify my LC method? Do you have any suggestion of suitable analytical column? Does anyone have same idea of what is happening and how to solve this?
I am getting crazy!
Thanks for help.