Headspace splitless residual solvents method
Posted: Wed Oct 03, 2007 9:57 am
I am having a problem choosing appropriate parameters for a residual solvents analysis using the headspace autosampler and a splitless GC method.
I have no choice about the basic parameters. We would not ordinarily choose to use a splitless headspace method for a residual solvents method we developed ourselves, but this method is being transferred from a different lab, and unfortunately has to be made to work substantially intact.
Here are the parameters:
Instruments:
Agilent G1888 Network Headspace Sampler
HP 6890 GC System
Column:
Phenomenex ZB-624, 30 m x 0.53 mm ID x 3 um film thickness; max T = 260 C (Isothemal/Program)
Solvents to be tested (level in standard in brackets):
MeOH (600 ppm), EtOH (1000 ppm), IPA (1000 ppm), CH2Cl2 (100 ppm), Methyl t-butyl ether (MTBE) (1000 ppm), DMF (200 ppm), Toluene (200 ppm), 1,4-Dioxane (35 ppm), 2-Methoxyethanol (a.k.a ethylene glycol monomethyl ether, EGME) (5 ppm).
This is made up in DMSO.
GC method:
Inlet: 170 C, 2.2 psi, Purge Flow 22.0 mL/min, Purge Time 0.60 min. Gas saver not on. Total Flow Setpoint = 27.2 mL/min
Oven: 40 C hold for 9 min. Increase to 200 C at a rate of 20 C/min, hold 6 min. Total = 23.00 min
Detector (FID): 220 C, H2 30 mL/min, Air 400 mL/min, Makeup (constant, N2) 40 mL/min, Lit Offset 1.0
Headspace method:
Oven: 140 C
Loop: 150 C
Transfer Line: 155 C
Vial Equilibration: 30 min
Pressurization: 0.50 min
Loop Fill: 0.50 min
Loop Equilibration: 0.50 min
Inject: 0.12 min
GC Cycle: 35 min
Carrier Pressure: 4.5 psi
Vial Pressure: 10 psi
Things I Don't Have a Choice About:
The following I would prefer to change only as a last resort because I need the method to be transferred essentially as-is:
The method has to be headspace, and unfortunately must be splitless. The inlet pressure and temperature are specified in the transferred method, as is the oven program and some of the headspace parameters. The temperatures chosen are those from the transferred method, as are the times. The column comes from the transferred method.
Things I Have A Choice About:
The lab that developed this method used different instrumentation, which I don't have immediately available to tell you but I can check if necessary. They didn't specify anything for parameters such as Purge Flow and Purge Time, and we had to make educated guesses about how some of their headspace parameters mapped to those used by the Agilent G1888, especially the times.
Observations:
It is necessary to adjust the Carrier Flow knob on the HS. Too high and the inlet pressure shoots up and cannot be maintained at 2.2 psi. I have tried to set the carrier flow to the maximum that will allows the 2.2 psi inlet pressure to be maintained.
The Problem:
I have managed to get the method to the point where I can see all of the peaks except for the 2-methoxyethanol peak in the standard. The transferring lab only gets a very tiny peak for this but it is below the limit of detection with the last method I used (I had 50 mL/min purge flow and 0.06 min purge time for that).
From that point I was trying to make adjustments, primarily to the purge flow and purge time, because these are parameters that weren't specified so I have the greatest leeway with, to get the sensitivity to the point where I can see the 2-methoxyethanol.
When I tried to run it this time though, I am getting no peaks at all. When I manually inject 1 uL of methanol I see a huge peak come out on the online plot, so I don't suspect anything of being wrong on my GC side. I found a post saying that the flow ought to be between 20 and 30 mL/min for a method similar to this, so I tried lowering my purge flow and increasing my purge time to the paramters above and still no peaks (except a tiny DMSO peak).
We don't have a lot of experience with splitless methods in our company, especially using headspace, and I wonder if anyone can offer any insights as to what I'm doing wrong.
Thanks!
Instruments:
HP 6890 GC System
Column:
GC method:
Headspace method:
Oven: 140 C
Loop: 150 C
Loop Fill: 0.50 min
Inject: 0.12 min
GC Cycle: 35 min
Observations:
The Problem: