Inert ion source vs Standard ion source on 5975C

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi!
A question concerning the difference in sensitivity between the Inert ion source and the standard ion source for Agilent MSD 5975C......

Is it true that inert source is more sensitive than std source?

Thank you!

Haven't noticed much of an increase in sensitivity, but it seems to go longer between source cleanings.