I believe my Chiralcel OF column may have been damaged through the use of ethanol. I was planning to rinse it with a 50/50 mixture of hexane/isopropanol and then test it with a chiral compound. My question is what would be a good test compound (and conditions) for this type of column?

Check with Chiral Technologies/Daicell and find out what they use to test the columns for QC.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

I contacted Chiral Technologies. For anyone else interested here are the test parameters:

Column Size: 0.46 cmI.D. x 25 cmL
Sample: trans-Stilbene oxide (1.0 mg/mL in mobile phase)
Injection: 10 uL
Mobile Phase: n-Hexane/2-Propanol = 90/10 (v/v)
Flow Rate: 1.0 mL/min.
Detection: UV 254 nm
Temperature: 25°C

For normal phase chiral application, TSO is standard compound for column efficiency, and you can use 1,3,5-Tri-tert-butylbenzene to measure dead time.
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 2 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry