USP on Mobile Phase
Posted: Tue Sep 25, 2007 6:56 pm
by shaun78
The USP suggests that, in order to make their assays work, one is allowed to adjust the mobile phase by 2% absolute or 30% relative, which ever is larger. However, the change in any component can not exceed 10% absolute, nor can the final concentration of any component be reduced to zero.
I just don't get it.
I have two bottles: a). water; b). acn.
Initial conditions: a). 82%; b). 18%
gradient over 20 min to: a). 40%; b). 60%
I want to increase the amount of water initially .... Would that mean I can only adjust the composition of, say my starting conditions, by 2%; or can I get away with 17% (can't reduce b to zero); or can I only run with 10 absolute variation%.
To me, I would be working in absolute, so I would think I can only change by 2% .... but we all know that really buys one very little if anything at all.
Like I said, I just don't get their terminology here.
Posted: Tue Sep 25, 2007 10:44 pm
by Mark Tracy
absolute: |%ACN(original) - %ACN(new)|
In your case 2% absolute means ACN(new) = 16% or
10% absolute means ACN(new) = 8%
relative: |%ACN(original) - %ACN(new)|/%ACN(original)
In your case 30% relative means ACN(new) = 12.6%
Clearly 30% relative is a greater change than 2% absolute, but less than 10% absolute. So you may change to as little as 12.6% ACN.
As a contrasting example suppose that your initial composition were 5% ACN. 2% absolute would mean ACN(New) = 3%, 10% absolute would mean ACN(New) = -5%, and 30% relative would mean ACN(New) = 3.5%. In this situation the greater change would be 2% absolute, and you would not hit the zero rule. So you could adjust to ACN(new) = 3%.
The third example would be 90% ACN initial. 2% absolute means 88%, 10% absolute means 81% and 30% relative means 54%. While 30% relative is greater than 2% absolute, it hits the limit of 10% absolute, so the maximum change would be to ACN(new) = 81%.
Posted: Wed Sep 26, 2007 12:13 pm
by shaun78
Mark,
Thank you for your exceptionally informative reply. I now think I understand what they are getting at with this.
Thank you!
Shaun
Posted: Wed Sep 26, 2007 7:41 pm
by Mark Tracy
I forgot one detail: these rules apply to every component of the mobile phase, so you need to do the math for both water and acetonitrile to determine the limits.
Posted: Thu Sep 27, 2007 11:16 am
by shaun78
will do.
Thanks again!
Shaun
Posted: Mon Nov 12, 2007 2:32 pm
by zokitano
Shaun78 has asked about mobile phase that contain two solvents (binary mixture).
What's happens when one deals with three solvents, for instance:
10% THF: 30%ACN:60% Buffer X?
Does it mean that you're allowed to do the same changes (in the specified limits given in Pharmacopoeia) for every solvent?
Or what about when one uses 4 different solvents?
Has anyone reasonable answer about this question?
I just wanna be sure about this.
Posted: Mon Nov 12, 2007 11:33 pm
by tom jupille
First of all, the suggested default adjustment limits appeared in Phamacopeial Forum in 2004. As far as I know, they have not yet appeared in the USP as such.
The FDA has similar limits for their use:
http://www.fda.gov/ora/science_ref/lm/v ... _04_05.pdf
This does cover ternary but not quaternary blends.
Posted: Tue Nov 13, 2007 7:20 am
by zokitano
Thank you Tom for the reference. It clarifies some of the doubts that I had.
Best regards
Posted: Tue Nov 13, 2007 12:18 pm
by shaun78
The USP brought it back to the PF around the time that I posted this question. Once again, it got shot down.
Because it did not become official, I had to quickly write a validation protocol and validate a method because I changed column length (150 mm to 250 mm) and mobile phase comp by 2%.
Incredible waste of time in my humble opinion.
Posted: Tue Nov 13, 2007 4:10 pm
by tom jupille
Yeah, but otherwise we'd be unemployed!
Seriously, if the situation is likely to recur, what I would do is to write an SOP essentially duplicating the FDA document (and referencing it for justification).